USP reference standards for peptide verification — anyone have experience?

raj_cambridge Sat, Jun 14, 2025 at 10:57 AM 10 replies 1,436 viewsPage 1 of 2
This thread is more than 11 months old. Information may be outdated. Consider searching for more recent discussions.
raj_cambridge
Member
489
2,123
Jun 2024
Cambridge, MA
Jun 14, 2025 at 10:57 AM#1

Two reports on the same product from two laboratories, four percentage points apart, and no method stated on either.

What I am after is whether a purity figure and a content figure can disagree without one of them being wrong.

If the honest answer is that nobody knows, that is a useful answer and I would rather have it.

42 12SleepFixSam, PurityPaulOR, MaxMetOK and 39 others
Reply Quote Save Share Report
RetaRick_CA
VIP Member
2,012
9,876
Jan 2024
California
Jun 14, 2025 at 11:07 AM#2

This one has a reasonably settled answer, so here it is. The provenance questions are more useful than the numbers. Does the report name the batch on your vial. Is the date consistent with when that batch existed. Is it addressed to the vendor or to the person who paid for the test. Was the sample taken from general stock or supplied by the vendor. A perfect certificate for a different batch is worth nothing, and vendor-supplied samples are the most common way a genuine report ends up misleading.

41 11AmyNC_wife, SkepticalSean, Dr.CardioMD and 38 others
Reply Quote Save Share Report
Dr.BariatricHTX
Senior Member
1,456
7,234
Feb 2024
Houston, TX
Jun 14, 2025 at 11:17 AM#3
RetaRick_CA said:
The provenance questions are more useful than the numbers.

Agreed, and mass spec answers a question purity does not: identity. High purity tells you one thing dominates the chromatogram, not that it is the molecule you ordered. A mass within about a dalton of expected is the identity check.

Correct me if the detail matters more than I have assumed.

40 10Dr.EM_Chicago, pete_RVA, CarlaRPh_TPA and 37 others
Reply Quote Save Share Report

Sigma-Aldrich — Research-Grade Standards

Certified reference materials, analytical reagents, and research-grade standards for peptide verification. Trusted by laboratories worldwide.

Shop Reference Standards
WendyG_ATL
Member
1,023
5,678
Apr 2024
Georgia
Jun 14, 2025 at 11:27 AM#4
raj_cambridge said:
Two reports on the same product from two laboratories, four percentage points apart, and no method stated on either.

This matches mine closely enough to be worth saying so. Inter-laboratory variation comes from real methodological differences, not sloppiness: which reference standard was used and how it was itself calibrated, the gradient and column, the detection wavelength, and where the integration baseline was drawn. A percentage point or two between competent labs is expected. What is not expected is a large gap, and the usual explanation there is that one report is not of the batch in your hand.

39 9AttorneyGrant, DebRD_ATL, KristenIndy and 36 others
Reply Quote Save Share Report
sophie_paris
Member
212
890
Nov 2024
Paris, FR
Jun 14, 2025 at 12:18 PM#5

Clinical perspective, offered as context rather than as advice.

Community-verified vendors for certificates of analysis: I maintain a spreadsheet of Janoshik-verified test results submitted by community members. Across 61 submitted tests:

  • 78% tested at ≥98% purity
  • 15% tested at 95-98% purity
  • 7% tested at <95% purity (concerning)

The vendors with consistently high purity results are listed in the Vendor Directory. We update monthly based on new community submissions.

38 8BethLabQueen, ChrisMacros, KetoKyle and 35 others
Reply Quote Save Share Report

Similar Threads

HPLC vs UPLC for peptide purity — method comparison study18 replies
Mass spectrometry for peptide identity verification — ESI-MS guide7 replies
Endotoxin testing methods — LAL vs recombinant Factor C17 replies
Red flags on COAs — how to spot a fake certificate5 replies
USP reference standards for peptide verification — sourcing guide3 replies
ForumsNewTrendingMembersAccount

Log In

Forgot password?
No account? Register